analogue output delays on PCI 6120

I have a VI performing magnetic measurements with a PCI 6259 and wanted to transfer it to using a PCI 6120 to take advantage of simultaneous sampling.  At its most basic the VI generates a waveform from a single output channel and captures 2 input channels (the output is then modified depending on the input but that is unimportant here).

When running on the PCI 6120 the analogue output seems to take an arbitrary amount of time to begin generating and then will not change when requested.

I have the generation and capture parts of the VI isolated in a small test VI (attached) which runs perfectly on the 6259 and has the problem described above on the 6120 if anyone thinks they can help.

Thanks
0 Kudos
Message 1 of 3
(2,975 Views)
Hi

I wonder if this KB article will be of any help to you

http://digital.ni.com/public.nsf/allkb/53B5101B1FBF017886256F3200748793?OpenDocument

Regards

YatinM
0 Kudos
Message 2 of 3
(2,952 Views)
Thanks for your help.  The problem was actually resolved by an NI engineer after a day or so investigation.  It seems that the 6120 (and 6150) have an extremely large internal buffer and in it default state the output will not update until the buffer is full so this buffer size had to be changed (please see engineers full response below if you are interested)

"I think I have found the root of the problem. The 6115 and the 6120 both
have a 30,000,000 sample buffer for output. By default all of this buffer
is used for generation. This means that new data written out takes a long
time to propergate through the buffer and actually be output. To overcome
this you can either, write data less often, output al lot faster or
decrease the buffer size.
To write data less often you just have to call the write function less, if
you are outputting at 50KS/s and writing 5000 samples with each write, put
software timing in the loop to make sure you only write 10 times a second.
To decrease the buffer size you will need to set the value of two
properties. These properties can be accessed from the DAQmx Buffer property
node as detailed in the attached picture. You will have to set them to much
smaller values and can set the onboard buffer to zero so you do not store
any smaples on the card."


I set both properties to the number of samples in my ouput waveform and everything works fine.