11-28-2005 05:18 AM
11-30-2005 04:43 AM
I don't understand why you connect permanent +5V to PFI0 because in that case you can't make a triggered acquisition by using PFI0 but anyway
Perhaps you can use PFI1 trig ext by declaring the good option in your AI start VI (trig digital option 2)
Good luck
jflab
11-30-2005 07:48 AM
Thank you so much for the answer.
There was a bad explanation maybe. I'm really using a trigger form +0V to +5V. The most important thing that I would like to know is: is correct the between PFI0 and the pin#11 (DGND) there are +5V without any external connection (at open circuit)
Thanks a lot
Moreno
11-30-2005 11:36 AM
OK I undestand
You need to put a pull-down of 1k or 4.7k beetween PFI0 and DGND (because of the internal pull-up at 5v of your DAQ card)
JFLAB
12-01-2005 03:25 AM
Thank you for solving my doubt!
Best regards!