 ksyz
			ksyz
			03-30-2006 10:11 AM
![Tom_W_[DE] Tom_W_[DE]](http://forums.ni.com/attachments/ni/130/2061/1/mean_robot.jpg)
		03-30-2006
		12:12 PM
 - last edited on
		05-07-2024
		08:35 AM
 by
		 Content Cleaner
			Content Cleaner
HI ksyz-
The M Series Help explains that the working voltage for your device is +-11V on any pin relative to AI GND. This means that if you exceed this limit on any pin then operation will no longer be predictable which explains why you're seeing strange results. I would suggest that you reduce the amplitude of your input if possible. You may need to add some external signal conditioning (such as an SCXI-1125) to meet this requirement.
Hopefully this helps-
03-31-2006 08:29 AM
03-31-2006 08:58 AM
Hi ksyz-
Yes, those are the only gain/range settings supported by your hardware. If you choose a range that is not directly supported, NI-DAQmx will automatically coerce the range to the next highest supported range.
Thanks-