Multifunction DAQ
SCXI-1104C
Got a running system with an SCXI-1104C module. I now have some measurement errors I never had before.
As a spare unit I have a SCXI-1104 (no "C")...  Is it direct compatible with the system ? Can I exchange
the 1104C for 1104 to eliminate multiplexor fault ?
//Beck777
(4,200 Views)
Accepted by Beck777
Hi Beck777,
You can find the manual for both the SCXI-1104 and the SCXI-1104, at the link below:
https://www.ni.com/docs/en-US/bundle/322147b/resource/322147b.pdf
If you read the common questions it explains that the only difference between the two is the cutoff frequency of the filters. It also says that  NI-DAQ version 6.7 or later is required for the SCXI-1104C whereas you only need NI-DAQ version 6.5.1 or later for the regular 1104. If you have the correct version of NI-DAQ it should work but be wary of the cut off frequency. You may end up with a lot of noise on your signal.
Tom T
Applications Engineer
National Instruments UK&Ireland
(4,177 Views)
