Read Dig Chan-Change Detection Event example does not work with USB 6221

"Read Dig Chan-Change Detection Event" example does not work with USB 6221. Namely: sometimes it reads wrong values in callback function. Also "Read Dig Chan-Change Detection" example does not work at all. It does not display any output at all.  
0 Kudos
Message 1 of 15
(4,685 Views)

Hi epertsov,

 

Can you please let me know what software and language you are using to program? Where is your signal coming from, and how fast is it? Have you verified your signals are correct by viewing them in a test panel when in Measurement and Automation explorer (MAX)? Does this happen with multiple USB ports? Posting a screenshot would be very helpful, so I could see what lines you are using as inputs.

Can you pinpoint the times when it does and does not work? I ask all of these because I was unable to reproduce the problem with a PCI card, and the 6221 is capable of doing this operation.

Message Edited by David L. on 05-10-2007 02:19 PM

David L.
Systems Engineering
National Instruments
0 Kudos
Message 2 of 15
(4,675 Views)
Hi David,
I use C++ example "Read Dig Chan-Change Detection Event" from "C:\Program Files\National Instruments\NI-DAQ\Examples\DAQmx ANSI C\Events\Signal\Change Detection". My signal comes to port0/line1 from a button, that I manipulate manually, so it is not fast. Here is what I got, when I just pressed and released button:
 
Data Read           Changed Lines
00000000             --------
01000000             -X------
00000000             -X------

The first line corresponds the event when I pressed button. The second line - when I released button. They are correct. The third (and the last) line is extra line that displays wrong value. I understand that this signal could be noisy, so it could be some extra lines, but I hoped that the last displayed value should be correct.

I also tried "Read Dig Chan-Change Detection" project from "..\NI-DAQ\Examples\DAQmx ANSI C\Digital\Read Values" directory and I couldn't get it work either.

Thanks and regards,

Eugene

 

 

0 Kudos
Message 3 of 15
(4,666 Views)

Eugene,

I ran the first example, and believe that you are seeing correct results. The Data posted is correct in that when you press the button, a change occurs, and is recognized on bit 1. When you release the button, another change occurs, and is again recognized. The reason why a change is marked for both the rising and falling edge is because the example configures the DAQmxCfgChangeDetectionTiming() for this.  You can change the function inputs if you would like it to act differently.

David L.
Systems Engineering
National Instruments
0 Kudos
Message 4 of 15
(4,651 Views)

Hi David, the problem is that in my connections when button is depressed, reading is 1, when button is released, reading is 0. I checked it in MAX. In this particular example when I pressed the button, I saw only one first line. when I released, I saw 2 new lines, last one with incorrect reading. It does not happen every time. But if you do this something like 1- 3 minutes with different speed, you alweas have this situation and sometimes you have not 2 readings per reconnection but 3,4,5...

What is related to "Read Dig Chan-Change Detection" it never comes to reading point at all.

Regards,

Eugene

0 Kudos
Message 5 of 15
(4,648 Views)

Sorry, I was mistaken in my previous replay: when button is depressed, reading should be 0, when button is released, reading should be 1.

Eugene

 

0 Kudos
Message 6 of 15
(4,647 Views)

Eugene,

I have personally run this example, along with LabVIEW examples, and cannot reproduce the issue you are seeing. One thing I did notice is that if the voltage is close to TTL Logic, than the value may change often and rapid. Please make sure that your values are 0-.8V for low and 2V-5V for high values. If there is any more detail that you can give me, that would be very helpful.

The “Read Dig Chan-Change Detection” never shows data because it is not coded too. It is coded to read the data forever( while (1) ) and does not have a printf statement in it. If you are not getting any errors, besides seeing no data, than the program is running. I may suggest either printing the data or replacing the while loop with a for loop, continuously building the data array, and then looking at it in the end.

David L.
Systems Engineering
National Instruments
0 Kudos
Message 7 of 15
(4,627 Views)

Hi David,

Our low level is about 0.9. I will make it below 0.8 and let you know how it works.

What is related to “Read Dig Chan-Change Detection”, I put there printing, but it does not print. Instead it gives me error -200284.

Thanks and regards,

Eugene

0 Kudos
Message 8 of 15
(4,613 Views)

David,

now we have good levels, but the same behavior. So the question is: are you working with 6221? and if so, maybe our device is bad and we should replace it?

Eugene

0 Kudos
Message 9 of 15
(4,604 Views)
I use LabVIEW rather than C++, so I may not be the most qualified to offer assistance, but I have been working with similar examples in LabVIEW recently and I also have a PCI-6221.  I think you were right in suspecting a noisy input signal since you are using a mechanical switch.  Have you looked at your signal with a scope?  I suspect you'll see several bounces on the signal before it settles down.  For troubleshooting, you could try using another digital line rather than the button to switch your signal and see if you still have the same problem.  This could also be the source of the error for the Read Dig Chan-Change Detection example.
 
If signal bounce is the problem, there are hardware filters you can get, or you could wait a few milliseconds after the initial change before reading the signal. 
 
 
0 Kudos
Message 10 of 15