03-20-2006 10:28 AM
I have the PCI-6221 DAQ with SCB68 set to continuous acquire at 1 sample at 1hz. I am acquiring voltage and current over time to monitor a battery charger, charging a battery. The system works fine on chargers referenced to ground. On floating chargers I get erroneous readings. I've tried adding a 10K bias resistor from negative terminal to AIGND as described in the manual. I've tried adding a 1M resistor from both positive and negative inputs to AIGND.
Any help I could get would be greatly appreciated
Darren
03-21-2006
12:46 PM
- last edited on
11-10-2025
11:20 AM
by
Content Cleaner
Hello Darren,
What kind of erroneous readings are you getting - noise or just incorrect voltages?
How are you measuring the signal - in Measurement and Automation Explorer or in LV? What input configuration are you using - NRSE, RSE, or Differential? Are you attaching the bias resistors to positions B and D as mentioned in the manual? You need to configure you measurement to be either non-referenced single-ended (NRSE) or differential when you create the NI-DAQmx task. The Field Wiring and Noise Considerations for Analog Signals is a good reference as you get started.
Good luck!
03-21-2006 01:21 PM
Micaela,
Thanks for the reply!
We have identified the problem to be with the voltage scale. I have a 10:1 divider on the voltage so I scale it as follows:
Pre-scaled value
Max Min
5.0 0
Scaled Value
Max Min
50.0 0
When I try and read 27 volts the value displayed is 10.62 volts. When I read 16.9 volts the value displayed is 10.62 volts. When I read anything below 10 volts the value displayed is correct. For now I'm just not scaling the voltage. What should I do to correct this?
Thanks
Darren
03-22-2006 06:30 AM - edited 03-22-2006 06:30 AM
If you read the specifications of analog Input of your hardware,
you will find that the maximum and minimum Voltage that the ADC of your card can measure is +10 V and - 10Volts respectively.
Anything above that, will show saturated +10 V reading on that analog input channel
You are speaking of an external 'voltage scale'??, are you not using this to scale down your analog input voltage to fall into the +/- 10V ADC range?
Message Edited by devchander on 03-22-2006 06:32 AM
03-22-2006 08:27 AM
I have a 10:1 voltage divider to measure the voltage. For example:
Battery voltage = 24V
Divided down = 2.4V
Now I apply my voltage scale to Map ranges
Pre-scaled value
Max Min
5.0V 0V
Scaled Value
Max Min
50V 0V
03-22-2006
08:39 AM
- last edited on
11-10-2025
11:21 AM
by
Content Cleaner
Hello Doliver,
Look at this KB and let me know if this is not the problem.
Message Edited by McKala[DE] on 03-22-2006 08:46 AM
03-22-2006 08:44 AM
I can't seem to open you're shortcut
Thx,
Darren
03-22-2006
08:47 AM
- last edited on
11-10-2025
11:21 AM
by
Content Cleaner
03-22-2006 09:18 AM
Thanks Micaela!
I was using the "Pre-scaled value" in the "Signal Input Rage"