01-19-2010 06:32 AM
Hello
I'm using PCI-6036E Card for data acqusition. We know this card has 16AI channels. Sampling rate rated is 250KS/s.
Suppose if i use all the 16 AI channels for data acquisition, then what might be the sampling rate of each AI channel?
Is it like 250 /16 = 15.6KS/s per each channel or each channel can acquire 250KS/s?
Thanks for your feedback
Regards
Kirit
01-19-2010 08:21 AM
Hi Kirit,
the device that you are using is an E-Series card that has one single ADC.
All channels are multiplexed to that ADC and you can have 200k AD conversions per second,
meaning that you will have 200/16 kS/s when reading all channels.
If you need simultaneous sampling you can go for an S-Series or some of the new X-Series cards.
Hope this helps.
David
01-19-2010 08:22 AM
Hi Kirit,
the device that you are using is an E-Series card that has one single ADC.
All channels are multiplexed to that ADC and you can have 200k AD conversions per second,
meaning that you will have 200/16 kS/s when reading all channels.
If you need simultaneous sampling you can go for an S-Series or some of the new X-Series cards.
Hope this helps.
David