Noise issue with USB6210 Daq

Hello,

    I'm using USB6210 to test a 8 channel signal with differential mode. The original signals are 8 channels photocurrent and then go to the same op-amps to be converted into voltages and then sent into the daq. I recorded the 8 DC signal for 10 seconds with sample rate of 30000. I found that the noise level of AI0 is alway higher than the other 7 channels. Is that normal? Is there any ways to reduce the AI0 noise?

Thanks,

Xiaotong

noise.jpg

0 Kudos
Message 1 of 3
(1,127 Views)

The best way to measure the noise of DAQ is to ground all the analog inputs and capture the signal.

Based on your description it is possible that your OPAMPs has a voltage offset.

 

If by grounding all the AI of DAQ results in almost comparable noise, then the issue is not in the DAQ but the external circuit, moreover the noise values are specified in the datasheet of DAQ but there is no guarantee that channel-channel matching is guaranteed.

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 3
(1,116 Views)

Also try swapping inputs of AI0 with another AI channel and see if the noise floor still follows AI0.

 

-AK2DM

~~~~~~~~~~~~~~~~~~~~~~~~~~
"It’s the questions that drive us.”
~~~~~~~~~~~~~~~~~~~~~~~~~~