10-18-2007 08:40 AM
10-19-2007 08:12 AM
Hey,
With an continous acquisition the "number of samples" declares the buffer size. If you don't set the number of samples the driver sets the size regarding your sample rate.
Christian
10-19-2007 08:21 AM - edited 10-19-2007 08:21 AM
If your acquisition is finite (sample mode on the Timing function/VI set to Finite Samples), NI-DAQmx allocates a buffer equal in size to the value of the samples per channel attribute/property. For example, if you specify samples per channel of 1,000 samples and your application uses two channels, the buffer size would be 2,000 samples. Thus, the buffer is exactly big enough to hold all the samples you want to acquire.
If the acquisition is continuous (sample mode on the Timing function/VI set to Continuous Samples), NI-DAQmx allocates a buffer equal in size to the value of the samples per channel attribute/property, unless that value is less than the value listed in the following table. If the value of the samples per channel attribute/property is less than the value in the table, NI-DAQmx uses the value in the table.
| Sample Rate | Buffer Size |
| no rate specified | 10 kS |
| 0–100 S/s | 1 kS |
| 100–10,000 S/s | 10 kS |
| 10,000–1,000,000 S/s | 100 kS |
| >1,000,000 S/s | 1 MS |
You can override the default buffer size by calling the Input Buffer Config function/VI.
NI-DAQmx does not create a buffer when the sample mode on the Timing function/VI is set to hardware-timed single point.
| Note Using very large buffers may result in diminished system performance due to excessive reading and writing between memory and the hard disk. Reducing the size of the buffer or adding more memory to the system can reduce the severity of these problems. |
For generations, the amount of data you write before starting a generation determines the size of the buffer. The first call to a Multiple Samples version of the Write function/VI creates a buffer and determines its size.
You also can use the Output Buffer Config function/VI to create an output buffer. If you use this function/VI, you must use it before writing any data.
The samples per channel attribute/property on the Timing function/VI does not determine the buffer size for output. Instead it is the total number of samples to generate. If n is your buffer size, setting samples per channel to 3×n generates the data in the buffer exactly three times. To generate the data exactly once, set samples per channel to n.
Regards,Message Edited by Neil S. on 10-19-2007 08:21 AM
10-19-2007 09:38 AM
10-19-2007 09:49 AM - edited 10-19-2007 09:49 AM
Hey,
As Neil allready wrote DAQmx is allocating the buffer....and LabVIEW does the rest.
When configuring a Task for continous acquisition the data will be written directly to your RAM via DMA. So the number of samples is the size of this buffer in your RAM. Then the DAQmx Read.vi takes the data from this buffer to your application memory. This memory will be allocated from the LV memory manager.
There are also other ways to set the buffer with the advanced DAQmx functions, and then only call the read VI within a loop if the number of samples are available. Attached you will find a little example on this (build in LV 8.2.1)
Christian
Message Edited by Christian_M on 10-19-2007 09:51 AM
10-19-2007 03:07 PM
Hello Redeemed,
To clarify your last post, the DAQmx timing VI handles the buffer acquisition. By passing a value to the number of samples, you are attempting to specify the buffer size. Depending on the sample rate, this value can be overwritten by the DAQmx driver. The Chart that Neil provided above indicates the values that the driver will set.
The DAQmx driver will automatically read the most recent samples from the buffer during a continuous acquisition. The position will also be automatically updated.
The DAQmx read property node includes all of the buffer control properties for DAQmx. The property Current Read Postion indicates in samples per channel the current position in the buffer.
10-22-2007 10:22 AM
Isn't the default behavior of the DAQmx driver to read the "earliest-acquired-but-not-yet-read" data from the buffer? I'm pretty sure it is. This supports the typical use -- repeated calls to DAQmx Read give you a lossless stream of data. You only have to call it often enough that it won't fill up and need to start overwriting.
(I've actually had apps where I needed to get "most recent samples", but I had to deliberately program around the defaults.)
-Kevin P.
10-22-2007 12:35 PM
Hello Kevin,
Thanks for your reply. I did misspeak in my earlier post.
By default, the DAQmx read will read the "earliest-acquired-but-not-yet-read" data from the buffer. Thank you for pointing this out.