Multifunction DAQ
NI9215 grounded differential or single ended signal measurement
Hello there,
I have a slight problem with my NI9215 - just to describe a situation, I'm measuring 4 channels of low voltage signal (lower than 100mV, mostly in range of 10-50mV) simultanously. At the moment I'm using scan engine to obtain values, will move to FPGA later to increase the acquisition rate. My signal is preconditioned with an amplifier, applying a gain of 1000 to bring it to mV range rather than uV. I'm using well shielded cables. I am not entirely clear whether my signal is grounded differential, or single ended - if I follow this manual:
https://www.ni.com/docs/en-US/bundle/ni-9215-getting-started/page/overview.html
Page 14 most closely resembles my circuit, thus I would assume it to be single ended. However, if I follow this manual:
https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/analog-fundamentals/grounding-c...
Figure 8, then in the table the right uppermost option describes my system best. Thus, I've tried two ways of connecting my sensors. First, as shown below in version A:
In this case, the noise levels are not so high, there however exists an offset between all 4 channels, even when measured voltage should be 0. Additionally, I cannot see the signal I am supposed to measure - if I make a secondary measurement with my Keithley 2001 multimeter (connected exactly as NI9215 to the system), I see a clean signal in the range of 10 mV - 40 mV and I can measure none with NI9215.
Then, I've tried the following connection scheme, version B:
The only difference is, I've added the connection between source ground and COM terminal - so to satisfy NI9215 manual, pg 14. In this arrangement, the offset error between channels is gone and I can see the right ampliude of the signal to be measured, but the noise levels are extremely high. I've tried adding a 1MOhm resistor in the line connection to COM, but it had no influence on the signal. I must say I am quite confused on how to resolve this situation. I would appreciate any advice on the matter.
Filip
(5,146 Views)
the module only supports +/- 10V, better accuracy can be achievef if you choose 9205 with a +/- 200 mV. Furthermore the way how you have to connect the wire depends on the sensor / source. On page - 12 14 you have to choose the right one. https://www.ni.com/docs/en-US/bundle/ni-9215-getting-started/page/overview.html
Is the noise inside the specification of the module?
Best Regards, Fabian
(5,049 Views)
Hi Fabian,
Thank you for your answer, however choosing the right wiring scheme (page 12 or 14) was actually part of my question - I am aware I have to choose the right one, but are unsure which one it should be - also the link to the manual is in my first post already - could you advise how can I make the selection?
At the moment using module 9205 is not possible for me. What I observe right now is that my amplifier circuit is adding a lot of noise to the system. I've moved to FPGA and the noise from shorted NI9215 module is around +/- 1mV. With added low-pass filter with 1kHz cutoff freuquency I can bring it down to +/-0.15 mV.
I was thinking about ungrounding my system, so that I deal with floating signal - that would correspond to figure 5 on page 13 from the NI9215 manual - which suggests to put 1MOhm resistor between minus cable and COM. However if we use the Wiring Considerations manual (https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...  ), the table 1, top row, left column - floating signal, differential measurement, the suggestion is to use two resistors - 10 to 100 kOhm to connect BOTH cables to ground (which I undestand represents COM port in this situation, which is then further connected to earth ground) - it seems to me that those two schemes are incomatible, so which one is the correct one?
(5,007 Views)