Hello,
I need to synchronize the data acquisition using 2 PCI 6143 cards, ie. the data acquisition has to start at the same time for the 2 cards.  Has anyone done this synchronization before using CVI?  Will appreciate your help.  Thank you.
0 Kudos
Message 1 of 9
(4,592 Views)
Hi!

   2 solutions, in my opinion.

   1) You can wire an external trigger signal to an analog trigger input, on both boards.  If you configure your acquisition to be triggered on this external signal, they will start together.

   2) RTSI cable. It is an internal bus, that connects directly the 2 boards.  That is, from PCI board 1 to PCI board 2.  Then you can use RTSI signals to synchronize your acquisition.

   Of course, I supposed the 2 boards are on the same PC. Can you specify your meassurement system configuration?  Maybe, it can also help to know what's your measurement purpose.

   Hope it helps!!!! Let me know.... Have a nice day!


graziano
0 Kudos
Message 2 of 9
(4,587 Views)

Thanks for your suggestion.  I have connected the two boards together using the RTSI cable. 

I have tried to set the source of the sampling clock to the first master device in DAQmxCfgSampClkTiming function call for the slave device.  However the outputs of the boards are not synchronized.  I'm not sure whether have I done it correctly.  Can anyone please advise?  Thank you. 

0 Kudos
Message 3 of 9
(4,571 Views)

Hi

Plz attach the code you have developed so that we can suggest better

 

0 Kudos
Message 4 of 9
(4,569 Views)
Hi!
   Have you configured the RTSI cable in MAX?

   I'm no longer working with RTSI cables, so I can't check from here...... but I remember RTSI cable should be configured in MAX.  In MAX help, just search for RTSI.

   Hope it helps!

graziano

0 Kudos
Message 5 of 9
(4,565 Views)
Hello!
 
To configure a RTSI cable in MAX, right-click NI-DAQmx Devices and Select Create a new NI-DAQmx Device...RTSI cable.
 
You can use the DAQmxExportSignal function to export the sample clock and the start trigger to the other device through the RTSI cable.

Message Edited by Michael P on 02-17-2006 11:53 AM

--
Michael P
National Instruments
0 Kudos
Message 6 of 9
(4,556 Views)
In this situation, you don't need to use the DAQmxExportSignal function. If you specify a terminal on one device when configuring another device, and these devices are connected with a RTSI cable as specified in MAX, NI-DAQmx will automatically route the signal over the RTSI bus.

geoff
--
Geoff Schmit
Huskie Robotics, FIRST Team 3061 Lead Mentor
http://team3061.org/
@team3061
Message 7 of 9
(4,546 Views)
Nice job, gcs!

0 Kudos
Message 8 of 9
(4,531 Views)

Hello,

Thanks a lot for your help.  Managed to do the synchroniztion.  I added the RTSI cable in MAX and added in the codes:             

\\ Dev 1

DAQmxErrChk (DAQmxCreateTask("",&taskOut1));
DAQmxErrChk (DAQmxCreateAIVoltageChan(taskOut1,"Dev1\ai0" ,"",DAQmx_Val_Cfg_Default,minIn,maxIn,DAQmx_Val_Volts,NULL));

DAQmxErrChk (DAQmxCfgSampClkTiming(taskOut1,"",rateOut,DAQmx_Val_Rising,DAQmx_Val_FiniteSamps,bufferSize));

\\ Dev 2

DAQmxErrChk (DAQmxCreateTask("",&taskOut2));
DAQmxErrChk (DAQmxCreateAIVoltageChan(taskOut2,"Dev2\ai0","",DAQmx_Val_Cfg_Default,minIn,maxIn,DAQmx_Val_Volts,NULL));

\\ Change the source of the sample clock here

DAQmxErrChk (DAQmxCfgSampClkTiming(taskOut2,"\Dev1\ai\SampleClock",rateOut,DAQmx_Val_Rising,DAQmx_Val_FiniteSamps,bufferSize));

Thank you once again.

 

 