09-17-2007 02:12 PM
Rob F,
Thanks for the help!
The product manual states the output is “TTL compatible pulse”.
Here is where I am at so far. See attached doc.
09-17-2007 02:35 PM - 編集済み 09-17-2007 02:35 PM
Message Edited by FlipTek on 09-17-2007 02:38 PM
09-17-2007 03:08 PM
Flip Tek,
Thanks for pointing out the missing reciprocal.
Question... are you using the Monarch SPS-5 Series infrared sensor?
After you drag the DAQ Assistant to the Block Diagram what setting are you using for your DAQ Setup?
Here's mine:
Express >> Input >> DAQ Assistant
Create New Express Task Window opens; Acquire Signals >> Counter Input >> Frequency
Supported Physical Channels opens; I select “ctr0”, It tells me to "Connect Your Signal to (Input Terminal): to PFI1" This is pins 16 and 1 on the NI 9401.
Configuration Window opens
Setting I select;
Acquisition Mode = Continuous Samples
Samples to Read = 100(default)
Starting Edge = Rising
Measurement Method = 1 Counter (Low Frequency)
Scaled Units Hz. I click run and 119 appears in the top Measured Values…
Team Stoney
09-17-2007 03:48 PM
Flip Tek,
I cannot seem to get counter input... any ideas?
Team Stoney
09-18-2007 08:49 AM
Hello Team Stoney and Flip Tek,
The reciprocal is only needed if your task is measuring the period of your signal. If your task is measuring frequency, then the driver will automatically make this calculation for you. If you are using the counters on a cDAQ, then you need to have your module in either slot 5 or 6.
119 (Hz or cycles/second) corresponds to the frequency of the measured waveform. In order to get RPM just multiple the frequency by 60. 119 (cycles/second) * 60 (seconds/minute) = 7140 (cycles/minute)
09-18-2007 09:07 AM
09-18-2007 03:51 PM
06-17-2010 12:24 PM
06-17-2010 12:49 PM
s-p wrote:
I am using M series USB-6229 DAQ board.i applied the same program ,but i can not getting reading of this vi.any help?
Do you have the same tach sensor?
06-17-2010 12:57 PM