Multifunction DAQ
Load cell amplifer signals
Hi All,
I'm trying to build an app to read a signal from a RS load cell & amplifier using a USB6210 daq card. I'v tried SE & Diff connections, and I see a change in signal when I apply a load, however it's not quite what I expected (see attached screenshots from MAX).
I'm supplying an external 5V to the amplifer, so with no load expect to see approx 3V from the amp - this works OK. With a load applied I expected to see a constant voltage drop, however I seem to be getting an unstable, spiked drop in voltage.
Any ideas why this is happening, any help would be much appreciated.
Martin
Link to RS load cell & amp spec sheet:
http://docs-europe.origin.electrocomponents.com/webdocs/0099/0900766b80099349.pdf
(4,267 Views)
Accepted by topic author Martin.Griffiths
Looks like a 100Hz hum ..... grounding and shielding and a clean powersupply for your amp??
Any switched power devices (powersupplies, converter, ....) nearby (shure your PC 😉 ?? Electronic ballasts for modern energy saving light or dimmer ??
Greetings from Germany
Henrik
LV since v3.1
(4,222 Views)
Thanks Henrik, using shielded cable solved the noise problem. I also changed the supply to the amp to give 12V instead of 5V, it seemed 5V was just below the operating range of the amp.
Thanks again.
Martin
(4,173 Views)
Waoooo good source of information for me.. Your progect was perfect indeed.
(3,455 Views)
