Hi M_caldon,
The card you have is actually a 6041E. This card does support analog triggering through the DIG_Trigger_Config command in Traditional DAQ. Here is the product manual that confirms this.
This was an issue with older versions of the driver, are you using the newest version of the Traditional DAQ driver?
Also, what input parameters are you passing to the DIG_Trigger_config funtion?
Thanks,
Abhinav T.
Applications Engineer