Multifunction DAQ
EMG signal analysis
I am using PCI 6040E, LabVIEW 8.6, BNC 2090 for the data acquisition. I want to analyze EMG signals for developing a prosthetic hand. I am new to this thing. Can any one please suggest:
1. Is my DAQ system alright to do these measurements?
2. What additional hardware do I require (Besides EMG surface electrodes, of course  for example do I require some amplifiers etc?
Many thanks.
(3,264 Views)
EMG can range up to about 10kHz, so your acquisition system has that covered.
I'm pretty sure you will need some instrumentation amplifiers to scale up the signal at least a factor of 10.  More importantly, it will give you some isolation between the electrodes and the acquisition system.
Randall Pursley
(3,261 Views)
You definitely need a medical-grade amplifier appropriate for acquiring EMG before the DAQ card.  You need both gain, safety (usually isolation, but at least low-leakage and protected inputs), an AC-coupled input (blocks DC offsets from electrodes and movement artifact), and filtering (to prevent aliasing of high-frequency noise).
You can find more information about this at our Biomedical User Group here:  http://decibel.ni.com/content/groups/biomedical-user-group
Look in the Wiki or Documents area - I posted a note about sources for amplifiers like these.
(3,243 Views)
