Data bandwidth
I'm looking at two Ethernet cdaq chassis connected together, one 4 slot cDAQ-9185 and one 8 slot cDAQ-9189. Was looking at populating them with NI-9234. My quick less educated crude calcs show the bandwidth should be enough to capture 48x channels (12 slots with 4 channels each module) @ 24 bit and 51.2kHz? Although I've likely made some mistakes.
(2,090 Views)
Accepted by PeterBrown
You're in the right ballpark for a raw calculation.  I'm assuming the 24-bit samples will be carried in a 32-bit integer though, which brings the bandwidth demand up to almost 80 Mbps.  If you're on a 100 Mbps connection, that may be too high a %age of the theoretical max to count on.  Others can probably comment with more complete knowledge.
-Kevin P
(2,074 Views)
Related article to calculate bandwidth - https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019QYVSA2&l=en-US
Santhosh
Soliton Technologies
Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution
(2,067 Views)
Thanks. That example in the link was for the same model of module so good to validate my rough calc
Which yes I only used 24 bit.. So explains why it was even more comfortable. I'll mostly be using them plugged into a gig ethernet port.
(2,035 Views)