Multifunction DAQ
DAQ-6008 Analog Input issues, non-zero voltage at pin when disconnected
I'm using the NI-DAQ 6008 to output a series of voltages and then reading a sense resistor using the analog input (RSE). I noticed that my analog input is giving me 1.3 V, when I probe it (relative to the device's ground), when completely disconnected. This alters the reading to give me a different measurement of the sense resistor than expected.
Why is my analog input pin non-zero??? Any help would be appreciated. Thanks!
(5,421 Views)
It's floating.  It could, in theory, be anything without something to somehow tie it to ground.  If you have it disconnected, don't read from it.
(5,419 Views)
That still doesn't explain things though.
The reading across my resistor using a voltmeter is 2.5V. When I make the same reading using the AI pin, it gives me 1.3 V.
(5,414 Views)
So? A dmm and a multiplexed daq card (with a pretty low input impedance) is not at all the same thing. Your reading is expected and you have been given the solution.
(5,411 Views)
Explain?
(5,407 Views)
Is one side of the resistor actually tied to ground?  If not, that is not a fair comparison.  And I thought we were talking about an open to the Analog Input.  Maybe pictures of your schematic would be helpful.
(5,404 Views)
Accepted by topic author catran
The 1.3 V is expected. The USB-6008 and -6009 have the world's strangest input circuit. The input impedance is about 144000 ohms terminated into 1.4 V. See the User Guide and Specifications document.
Lynn
(5,400 Views)
