01-30-2006 05:50 PM
01-31-2006 12:02 AM
For what do you want to configure a load cell?
Do you want to measure a load cell using a USB 6009??
A load cell that is connected to a suitable Excitation circuit, gives a variation in its voltage output wharever a load is applied.
All you have to do is measure this variation in voltage on your Analog input channel( First check if this voltage is within your DAQ card's input range limits), scale it using any relevant conversion techinque and display load in its engineering units.
if you need any further clarifications, ask in detail
Regards
Dev
01-31-2006 01:19 PM
01-31-2006 01:23 PM
02-01-2006 04:15 PM
02-02-2006 02:19 AM
Hi,
Yes it would indeed be a good idea to go for signal conditioning.
I guess SCXI 1520 ( it also sources excitaiton to the sensor)should be applicable .
But first, reads its specs and see if it can support your load cell measurements.
Also you can do load measurement using your existing DAQ card without a signal conditioning module.
I once measured a Load cell ( Note: was getting its excitation voltage from an independent source) using a DAQ card without using any signal conditioning module.
First, i amplified the load cell output, by using a highly accurate external isolation amplifier.
In my case, i brought it up to +/- 10 volts range and then, used the AI channel on a E seried DAQ card to measure the voltage and derive the load using a Look up table.
Hope this info is of help to you
regards
Dev