Multifunction DAQ
Cannot measure AC current using NI VirtualBench - NI VB-8012
Hi All,
I’m trying to perform manual AC current measurements, but a result is just a noise.
I have a simple series circuit with FGEN, as VirtualBench, and load resistor of 50[Ohm]. My measuring setup is as described in the attached image. Except the batteries – I use fGEN at frequency of 486kHZ.
I’m pretty sure that my measuring setup is correct, because when I switch to DC mode – I have a comprehensive current measurements. In addition, when running in AC mode, through the AC multi meter connection, the load resistor is warming up – which means that there is a current in my circuit.
Both fuses were tested for continuity – and were fine.
What do I miss here? Is there some kind of “reset” for the virtual bench?
Thanks in advance
Art
(3,543 Views)
Accepted by topic author ArtB83
I have a simple series circuit with FGEN, as VirtualBench, and load resistor of 50[Ohm]. My measuring setup is as described in the attached image. Except the batteries – I use fGEN at frequency of 486 kHz.
I'm not an analog engineer, but I hypothesize that this circuit's frequency is beyond what the VirtualBench DMM can measure.
Looking that the specs, it appears that the DMM has guaranteed accuracy for AC signals up to 5 kHz:
If you reduce your circuit's frequency to 5 kHz or lower, do you see good measurements?
Joe Friedchicken
NI Configuration Based Software
Get with your fellow OS users
[ Linux ] [ macOS ]
Principal Software Engineer :: Configuration Based Software
Senior Software Engineer :: Multifunction Instruments Applications Group (until May 2018)
Software Engineer :: Measurements RLP Group (until Mar 2014)
Applications Engineer :: High Speed Product Group (until Sep 2008)
(3,517 Views)
Hi Joe,
You were 100% right - When I dropped my work frequency to 5kHz - the ampere-meter started to show correct values.
Thank you very much.
Art
(3,507 Views)