Hello chenchie,
There are two types of calibration that can be performed on your DAQ device: internal or self-calibration, and external. An internal calibration adjusts the calibration coefficients relative to the internal reference onboard the DAQ device. The internal calibration requires no external connections or calibration equipment, and can be performed in Measurement & Automation Explorer (MAX), by right-clicking on your device under Devices and Interfaces > NI-DAQmx Devices, and selecting Self-Calibrate. An external calibration adjusts the calibration coefficients of the device relative to a high-precision voltage source external to the device, and can be traceable to NIST (National Institue of Standards and Technology) standards. National Instruments provides services for having your device externally calibrated. More information can be found at www.ni.com/calibrate.
Also, more details about internal and external calibration for your device can be found in the E-Series Help Manual.
I hope this helps,
Travis G.
Applications Engineering
National Instruments
www.ni.com/support
07-01-2006 04:03 PM
Hi Travis,
Thanks a lot for your information. It successifully solved one of my PCI 6024E DAQ card problems.
I have two PCI 6024E, and I use them to measure temperature with K type thermocouple. One of them has a 6 degree C offset and has been solved by your method. But other one has a -100 degree C offset, and it fail to calibrate by self-calibration. Is there any other method to fix it?
Thanks a lot,
Chen