 Forrest_Bao
			Forrest_Bao
			03-10-2009 12:06 AM
Hi, I am facing an application on sampling EEG signals. The amplitude is in the range of 1 uV to 500uV. The accuracy is required to be 1uV. I checked several high-resolution DAQ card. But their range accuracy in minimum voltage range at least 10 times (like 28uV for PCI-6284) larger than required accuracy. So, I want to amplify the signal by at least 100 times.
Is there any NI hardware or build-in amplifier in the DAQ card that can amplify the signal by 100 times? And, the larger, the better.
 kikiduu
			kikiduu
		03-10-2009
		01:15 AM
 - last edited on
		04-25-2024
		09:20 AM
 by
		 Content Cleaner
			Content Cleaner
It is not just enough to divide the voltage range by resolution to get the minimum voltage measurable. You should look into the noise and accurancy specifications of the card also. I would have suggested some SCXI modules, but the minimum system noise is 20uV in some modules. So why not try 5922 Digital Oscilloscope.
 Steve.J
			Steve.J
			03-10-2009 09:07 AM
PLEASE do not attempt to acquire signals directly using a NI DAQ board or any NI signal conditioning products. NI does not make any hardware that is suitable for direct connection to ANY living organism. There are two big problems:
1. SAFETY - you need an isolated amplifier system with appropriate safety certifications (UL 60601-1 or EN60601-1) which is critical to provide protection against shocks and/or burns in the case of an electrical fault. And not just faults originating from the EEG system - also protection from providing a ground path for electrical faults originating from some other device that may come into contact with the subject. If you are trying to record from a frog or something, you may not care so much about safety, but then you have the next problem:
2. PERFORMANCE - surface biosignals (EEG, EMG, ECG, etc.) require specialized amplifiers to acquire cleanly. The signals are riding on top of 56/60Hz line frequency interference that is often orders of magnitude greater than the actual biosignal. The signals are also riding on a large DC offset generated by the electrodes and electrolyte used to make a good connection with the skin, and the source impedance (electrode to skin/body) is high (5-10Kohms or more) and often badly mismatched. Signal conditioners for biosignals are designed specially to deal with these issues - they are usually AC coupled, have adjustable low and high pass filters (and often a notch filter at the line frequency) to block the offset potential and prevent noise from being digitized and aliased, have multiple stages of gain to reduce amplifier blocking (saturation) due to movement and other transients, and have very high impedance balanced inputs (high CMRR) to reject the common-mode noise and only amplify the signal of interest.
Specialized amplifiers are available from a number of sources and have high-level analog outputs that are ready to be digitized by an NI DAQ board. Look at companies like Grass Technologies, AD Instruments, MIndware Technologies, and others.
Steve