12-06-2005 04:59 AM
12-06-2005 05:28 AM
Hi Manuj,
When you refer to Speed of 1.25 MSamples/sec, What you speak about is the maximum sampling rate the input circuitry can digitize the analog signal or the card can sample sample that many individual samples in a second.
if you use 1 channel out of 8 channels in the 6070, you can go to a max sampling rate of 1.25 Msamples /sec
If you are using 8 channels also you can go fro a sampling rate of 1.25Msamples/8 = approx 156.2 kilosamples/sec( Ignoring the settling time considerations)
Where as, Resolution is the number of bits that the ADC uses to represent the analog signal.
The higher the resolution, the greater the number of divisions the range is broken into, and therefore, the smaller the detectable voltage change for a given input range and gain. The smallest detectable voltage change for an ideal DAQ board is determined by: Smallest detectable voltage change = range/(gain*2resolution)
If you select the ADC range as +/_10 V, since your card has 12 bit resolution, The ADC is scaled to represent +/-10 V in 4096 steps where each step= 4.8 mV
Hence the 12 bit resolution determines how accurately you can measure an input signal.
Hence, they both are independent characterstics and do not correlate in the way you have understood it.
Hope this answers your query
Regards
Dev