Hi,
please have a look on the attached VI. I have test it and it works fine.
Best regards
Rüdiger
06-25-2008 05:38 AM
Hi Rüdiger,
thanks a lot, but it doesn't work. I always get error -200016.
Maybe it is something with the PC. I run an Lenovo ThinkCentre MT-M 8811-CTO (3.4 GHz Pentium D, 1GB RAM).
Best regards,
Michael
06-25-2008 06:09 AM
Hi,
you should reset the device in the Measurement & Automation Explorer and run the selftest.
Try to generate a sine wave with 1MHz rate for one channel in the testpanel. Is that working?
You should create a task in Max including all ao-channels and run it (see attachement). Is that working?
What version of Windows is are you using?
What DAQmx driver version are you using?
Is it possible that you test the device on another?
Best regards
Rüdiger
06-27-2008 01:37 AM
Hi Ruediger,
I tried to generate a 1MHz Sinus out of the MAX, but it does not work. I get the same error as in my application. The limit is around 45 kHz. I tried it on another system we are running with the same IO-Card and PC Hardware. There it also does not work.
We are using Windows XP Professional with SP 2.
The DAQmx driver Version is 8.5.0f3
Thanks a lot and best regards.
Michael
06-27-2008 01:59 AM
Hi,
I would recommend to update the DAQmx driver to version 8.7.1.
Link:
http://joule.ni.com/nidu/cds/view/p/id/965/lang/de
Please test the device again after the driver update. It should work with one channel and a sampling rate of 1MHz.
Best reagards
Rüdiger