Machine Vision
Image Prespective Error
The image of part to be inspected using camera and product in perpendicular direction.
1. Basler Camera :Basler acA2440-20gm
Sensor Format 2/3''
  Sensor Type CMOS
  Sensor Size 8.4 mm x 7.1 mm
  Resolution (HxV) 2448 px x 2048 px
  Resolution 5 MP
  Pixel Size (H x V) 3.45 µm x 3.45 µm
  Frame Rate 23 fps
  Mono/Color Mono
2. Lense  : Computar M1214-MP2  (f=12mm F1.4)
But the image seems to be not correct towards the right end as if dilated some perspective / parallax/distortion. The image is finally used to do geometric pattern matching.
Please help to fix it or any other additional set up used.
(1,581 Views)
This looks like a case where I would seriously consider a telecentric lens.  You are going to have some parallax errors with the current lens no matter what you do.  If you move the camera back and use a longer focal length lens, it would reduce but not eliminate the parallax.  A telecentric lens is the only way to completely eliminate parallax.
Bruce
Bruce Ammons
Ammons Engineering
(1,554 Views)
Bruce is correct that telecentric lens is the best. But the other option of moving the camera away from the object and using longer focal lens will help a lot. Should be sufficient for geometric pattern finder.
Try to use f = 45 or similar.
Other option is to use image calibration to correct the image before you do the processing.
https://zone.ni.com/reference/en-XX/help/370281AG-01/nivisionlvbasics/calibrating_images/
https://zone.ni.com/reference/en-XX/help/370281AG-01/imaqvision/imaq_correct_calibrated_image/
Good luck
Amit
Amit Shachaf