03-24-2006 02:07 AM
The attached code will give me 52 ms at the Start_Voltage and 52 ms at the Stop_Voltage.
With the Step_Voltage = 0 the voltage will only switch between the two voltages.
I tried to add code from the 1ms Delay discussion but was unable to get the code to compile (a lot to do with not understanding what was going on!!)
Did someone get it to work and if so would they be willing to help?
Can I reduce this time down to 1 ms or so?
Thanks for the help
Simon
03-24-2006 06:39 AM
This is the file, sorry.
Simon
03-24-2006 07:54 AM
hpe363xa_ConfigureVoltageLevel (PSUHandle, "1", Stop_Voltage);
Loop_Count--;
03-24-2006 08:44 AM
I have a GPIB interface so I belive it is up to 1ms switch as is the Agilent E3632A power supply.
I am not waiting for a response back from the unit should also help!
Yes, no delay between Stat_Voltage (12V) and Stop_Voltage (10V) while Loop_Count will define the number of times you switch between the two voltages.
52ms at 12V then 52ms at 10V, the rate of change between the voltages is a function of the PSU (dose that sound right??!! )
I have two different PC's with different processors and operating systems but still at 52ms on each.
Is there a way to check what time it takes to execute the two instructions?
If this is less that what I am seeing from the output from the power supply then maby I will have to live with 52ms!!
Thanks for the input
Simon
03-24-2006 09:05 AM - edited 03-24-2006 09:05 AM
If you look at the specification for Agilent E363xA series power suppliers you'll see Command processing time < 100 msec, where this time is intended as "Maximum time for output to change after receipt of commands".
According to these figures, your 52 msec is probably the best performance you can expect from your supplier.
This can explain why you aren't seeing different figures using the supplier with different compures: the bottleneck is the supplier itself and you won't be able to shorten it.
Message Edited by Roberto Bozzolo on 03-24-2006 04:05 PM
03-24-2006 09:21 AM
Roberto, you could be right, Thanks.
Has anyone got a power supply to do fast transients (1 ms) under program control !!!?
Thanks for the help
Simon
03-24-2006 09:50 AM - edited 03-24-2006 09:50 AM
The type of equipment to get to 1mS resolution will depend on the power output. For low power you could use an analog DAC card to drive the waveform and then run it through a suitable amplifier. This allows you to preload the waveform onto the card (provided the card is capable of this) and then trigger the output pattern. For something with a little more output juice you could look at one of the source meters from keithley, something like a 2601. These are dc sources that can be programmed to execute programs internally to output the desired voltage waveform. The catch is that you really have to pay attention to the output connection to make sure the capacitance does not over run your output step rate.
The bottom line is that a windows PC itself is going to be too slow to drive the waveform you are looking for. You will need to preload it into a hardware buffer like a DAC card or use some type of dedicated embedded controller in the hardware to execute the waveform.
Message Edited by mvr on 03-24-2006 09:51 AM