For your first question: In your specific VI you are using the Open_Create_Replace VI. In one case you're using it with the "open or create" mode. In the other case you're using it with the "create or replace" mode. This second way is what's causing the advisory dialog. If you don't want to display advisory and replace the file anyway, then wire a True constant to the "advisory dialog?" input.
For you second question: The 6255 does have a certain amount of memory, and it's limited to 4095 samples. How this memory is used is dependent on how you do the data acquisition. If you do buffered acquisition then you're basically having the card collect data and store it into its buffer. Then your application would read data out of the buffer. If you don't read data out of the buffer fast enough then you will get the buffer filled up and then new data will have to force out old data. There are lots of examples that ship with LabVIEW that show you how to do data acquisition. Open the Example Finder (Help->Find Examples, and in the tree open "Hardware Input and Output" and then "DAQmx". You can also start here: Getting Started with NI-DAQmx.