Did you mean: 

ni max is missing USB device

Device manager shows the USB Devcie and connected, but it is missing in NI MAX. this is in windows 10, Labview 2020, 64 bit.

0 Kudos
Message 1 of 8
(1,526 Views)

You probably don't have the LabVIEW drivers needed to communicate with the device. What device are you working with? (manufacturer / model)

0 Kudos
Message 2 of 8
(1,493 Views)

Equipment :Keysight U2000 Power sensor thru USB

0 Kudos
Message 3 of 8
(1,454 Views)

@jbb3910 wrote:

Equipment :Keysight U2000 Power sensor thru USB


Does the USB connection to the device appear as a virtual Com port?

 

 

========================
=== Engineer Ambiguously ===
========================
0 Kudos
Message 4 of 8
(1,442 Views)

not . It appears as a USB Test and Measurement Device (IVI) in device manager.

0 Kudos
Message 5 of 8
(1,440 Views)

Do you have NI-VISA installed? Which version?

0 Kudos
Message 6 of 8
(1,423 Views)

ni-visa2020 on Windows 10 PC

0 Kudos
Message 7 of 8
(1,411 Views)

You may need to install Keysight drivers, have you installed those? From a quick search, it looks like you might need:

 

IO Libraries Suite : 

  • Enables instrument communication for a variety of development environments (MathWorks MATLAB, Python, NI LabVIEW, Microsoft Visual Studio and more).

RF Power Meter Drivers 

0 Kudos
Message 8 of 8
(1,403 Views)