01-19-2012 02:59 PM
I just installed LV 2011 and all necessary drivers and using MAX to ping & find 3 connected GPIB instruments. It finds 2 of them but the third one it says is Agilent N1913A, a power meter, which is NOT connected, but WAS connected long before I installed Labview & Max. It does NOT find the Agilent N5230A PNA connected to the GPIB bus. Any ideas why this is happening?
01-19-2012 09:10 PM
Have you run a refresh in Max?
01-20-2012 05:45 AM
Yes. Often.
01-20-2012 05:56 AM
May be you have to select 'GPIB' communication option in the instrument itself & also check address may be it clashing with other GPIB complaint instrument.
01-20-2012 06:06 AM
I have GPIB selected and a different address from everything else.
01-20-2012 06:19 AM
Try to swap the GPIB cables with the successful connected instrument and this one. Might be cable is faulty.
In which configuration you have connected all gpib instruments.
01-20-2012 07:15 AM
I found the problem. Refresh doesn't do it. I had to "Scan for Instruments" and it found it ok. Thank you for your help.
01-20-2012 07:22 AM
Thats not a problem as I see it.