07-03-2009 11:40 AM
Hello
I have a doubt. I'm using NI daq assistant to aquire a signal from a device. I'm reading 6 channels simultaneously and continuously, at 1kHz.
However, I checked that the loop where the whole process runs makes 10 iterations in 1second , which give me 6 channels x 1 sample/channel, 60 samples in a second. (or 10 samples/channel)
This is kind of far from the 1kHz that I expected. The loop has no delays, nothing related with time... Am I confusing everything, missing something, or my computer cannot process teh cycle faster that 10 Hz?
Please Help...
Thanks in advance
Solved! Go to Solution.
07-03-2009 12:03 PM - edited 07-03-2009 12:04 PM
I suppose you use the DAQmx Read vi to read samples.
What did you specify at the connector “number of samples per channel”?
With a scan rate of 1kHz and a delay time of 0.1s I would say 100 Is this correct?
Are you sure you read all available data from the buffer?
I mean, it’s pointless reading every sample one by one.
So I don’t see where your “1 sample/channel” come from
Can you post the vi you are using?
It will help us a lot in finding the error.
Regards
07-03-2009 02:59 PM
You also might want to take a look at the task set up in MAX.
If you are set for N Samples, then you will only get the number of samples specified when you read.
If you are set for Continuous Samples, the system will continually put data into the buffer.
We generally use Continuous Samples in our systems and create the number of samples (which becomes your buffer size) for larger than the number of samples that we expect to read. For example we have a system that samples at 1kHz and we read all of the available data 10 times a second. We still set our number of samples to 200. We have plenty of time to get through our loop and read more samples.
Those are just some thoughts on setting up tasks for DAQmx.
Rob
07-06-2009 03:48 AM
Hello
First, I would like to thank for the replies.
At DAQ assistant I set continuous samples and a sample frequency of 1 kHz. I want to have a program that allows me to give information in real time and allows me to make the average and standard deviation each 2 seconds. I expected to have 2000 samples to average (if the iteration time of the loop was 0.001 s (or 1/1kHz)), but I only have 20 samples in 2 seconds.
Thus, how can I improve the time spent at each iteration?
Hope my question is clear and thanks again for the help
Best regards
07-06-2009 04:19 AM
It's not because you have a sampling frequency of 1kHz that your iteration loop in LabView needs to run at 0.001s
The buffer on the DAQ will capture x samples and with one "Read DAQmx" instruction you read the x samples at once.
If you setup your DAQmx task to read 2000 samples "Samples to Read" in "Configuration" tab or wire 2000 to connector "number of samples per channel" of the "DAQmx Read" vi, that vi will wait for 2 seconds.
You can then average those 2000 readings with the "Mean" vi.
Hope this helped 🙂
07-06-2009 05:35 AM
Hmmm...ok! Thanks Alain S. It sure help!!
Unfortunately I don't have access to the DAQ board in the next 2 or 3 days, but when I try what you said I'll let you know
Thanks for the replies and for the help
07-08-2009 03:19 AM
Hello Alain S
You were absolutely right!
I changed the number of samples and the number of iterations per second icreased considerably, allowing me to do want I wanted.
Thanks for the help