display of 16-bit images

Several weeks ago, I was having a problem displaying images from two different framegrabbers at the same time. It finally came down to installing the latest IMAQ drivers to fix the problem. Unfortunately, this seems to have caused another problem to occur. None of the images I'm displaying on the screen look quite right anymore. If I move my mouse over the image, it will momentarily "flicker" and look normal for a moment, but then right back to looking wrong.
0 Kudos
Message 1 of 3
(2,456 Views)
Hello JKLambert,

Do you acquire correct images from each camera separately when doing a Grab in the Measurement & Automation Explorer (MAX)?

Can you run two instances of the HL Grab example from the NI Example Finder at the same time to acquire correct images?  (Make sure the image name input of the IMAQ Create function is different in each instance.  Acquire from a different framegrabber with each HL Grab vi).

Allen H.

0 Kudos
Message 2 of 3
(2,438 Views)
I got it working! Turns out I had to change my image displays from full dynamic range to a set range. This was not necessary before I upgrade IMAQ drivers.
0 Kudos
Message 3 of 3