difficulties connecting 2 bnc-2110 signals to the scope
maybe someone has encountered the same difficulties...
I have a circuit with DAQmx blocks that generates a waveform and writes it to the buffer.
The outputs channels 0,1 concide with the outputs on the bnc-2110 block.
These outputs are wired with coax to scope channels 1 (x) and 2 (y).
Now when I disconnect 1 cable I see a perfect waveform. But once I connect both cables to the scope
I get only a perfect reading at channel Y. Channel X is corrupted.
When I switch the cables at the bnc-2110 and leave the connection at the scope, still Y is perfect, X is corrupted.
So connecting output channels 0 or 1 to scope channel 1 (x)  does not make a difference. X stays corrupted.
The same holds for Y. It stays perfect.
The odd thing is that when I connect scope channel 1 (x) with a waveform generator, and scope channel 2 (y) with
channel out 0 or 1 of the bnc, I get 2 perfect readings. Only thing I can see is that the output of the generator is grounded.
When I reverse this. I get a perfect reading from the bnc at scope channel 2 (y) and again a corrupted signal at scope channel 1 (x)?!
So it seems that I get 2 perfect readings when x from the bnc has been replaced by the function generator.
Both should be grounded. The switch is in position GS (instead of FS). I connected the scope and computer to the same
electrical adapter, to be sure, but still have the impression I have grounding issues...
Bob
(2,531 Views)
Hi Bob,
Sorry for the late answer.Is your problem solved?
If no, could you be a little bit more specific when you says DAQmx blocks? Could you say us which hardware you are using and what is your program?
So we can do the test or try to reproduce the problem.
Julien Roland - District Sales Manager
NI Belgium - Technical Support
Don't forget to rate a good answer
(2,489 Views)
