Jarrod,
thank you for your comments.
the subvi does not seem to have custom scaling inputs. The inputs min and max to set the bridge range seems to be the issue. the scale of -0.01 to 0.01 is out of range according to the error, but this is a resonable value for strain. I have attached an *.LLB file which contains the subvis used in the program. I have also attached a modefied version of the program. It still has the problems with shunt callibration, I only changed the sample clock mode and read method.
thanks for your help.
Alfonso
01-17-2007 05:25 PM - edited 01-17-2007 05:25 PM
Message Edited by Jarrod B. on 01-17-2007 05:25 PM
01-18-2007 10:50 AM
Jerrod,
thanks for your reply, it was usefull to go back to the original example program. I still had the error, but tracked it down to a hardware problem since you did not have any issues running the example program. one of the NI 9945 bridge completion modules had a bad screw terminal. I identified one module before starting that was bad but this particular module seems to have a loose screw which does not allow for good connection of the strain guage terminal, it was having intermittent contact. I did not recognize it before as I tested the guages using MAX before hand and everything was OK.
I guess I received a bad lot of bridge completion modules.
Thanks, problem solved.