ni.com is currently undergoing scheduled maintenance.

Some services may be unavailable at this time. Please contact us for help or try again later.

LabVIEW

cancel
Showing results for 
Search instead for 
Did you mean: 

daq question

I would like labview to read the temperature of a omron digital panel meter
with a 3 wire pt100 sensor. I don't want to diconnect the sensor from the
existing meter. How can I read the resistance of the sensor without reading
the input res. from the meter?

thanks,
w
0 Kudos
Message 1 of 3
(2,680 Views)
First off, you must ensure that the daq channel you wish to use is a high impedance input. This is not normally a problem, as any daq board I have used is in the hi-K's to Meg ohms input.
The Omron will normally supply a constant current drive to the RTD, and then measure the voltage across it, internally in the meter. You could hook your daq input across the RTD resistance terminals ( not the lead compensation terminal), and measure the voltage. You would then have to do a conversion to engineering units (temperature) in Labview. You will need a table of resistance vs. temperature, and ohm's law to do this.
0 Kudos
Message 2 of 3
(2,680 Views)
Thanks,

w.


"Process Systems" wrote in message
news:506500000005000000147B0100-1079395200000@exchange.ni.com...
> First off, you must ensure that the daq channel you wish to use is a
> high impedance input. This is not normally a problem, as any daq board
> I have used is in the hi-K's to Meg ohms input.
> The Omron will normally supply a constant current drive to the RTD,
> and then measure the voltage across it, internally in the meter. You
> could hook your daq input across the RTD resistance terminals ( not
> the lead compensation terminal), and measure the voltage. You would
> then have to do a conversion to engineering units (temperature) in
> Labview. You will need a table of resistance vs. temperature, and
> ohm's law to do this.
0 Kudos
Message 3 of 3
(2,680 Views)