What is a fast way to drive a HP53132A counter to do real time, "time interval" measurements?
Right now I run my system on a simple PCI GPIB bus. The fastest I can the time interval measurement is at 100 Hz. However, the has 350 ns timing and the PC running the program is much faster than this. So I should be able to achieve a rate of at least 10 kHz (depending on the counter of course). Do you have any suggestions for the way to set up the counter for the fastest measurement?
(2,768 Views)
I think that you're running into a limitation of the counter itself. My manual for the 53131 says that it can can do a maximum of 200 GPIB measurements per second. I think that the 53132 is similar except for extended resolution. What you might try is the STATS mode where you command the counter to take a number of measurements and then calculate the statistic. It is generally fster to let an instrument do a series of measurements and accumulate the results than to try and transfer one measurement at a time over the GPIB bus. VXI, PCI, PXI instruments can do things a lot faster so that might be an option as well. You should also be aware that your rate of 10 kHz requires timing resolution greater than what is in a PC (a PC has 1 msec resolution at best a
nd it's not deterministic at all) so if you need something accurate, you'll have to use external hardware timing.
(2,768 Views)
