VISA comport wrong data
Thinking this might work. Can't check it now because the equipment is running.
Message 11 of 19
(1,895 Views)
You must put a delay between the property node and the Visa read function. Without this, the buffer won't have time to receive datas.
“En science, la phrase la plus excitante que l'on peut entendre, celle qui annonce des nouvelles découvertes, ce n'est pas "Eureka" mais c'est "drôle"
Isaac ASIMOV
Message 12 of 19
(1,891 Views)
No. Not true.  Typically you would see a delay between a VISA Write and a Bytes at Port to give time for a response.  No need for a delay between Bytes at Port and a VISA Read.
The correct solution to this problem is what Crossrulz stated.  DON'T use bytes a port, just use a very large number at the VISA Read and allow the termination character that is present in the message and has been enabled at the Serial Port Configure to do its job.
Message 13 of 19
(1,887 Views)
One thing you can do: Monitor the output in raw mode to see if the /03 characters really appear or not.
I add a Q&D monitor vi I once wrote. Configure the rs232, gnore the modem-line settings, run it and look at the data comming in 😉
Greetings from Germany
Henrik
LV since v3.1
Message 14 of 19
(1,879 Views)
Here is my new corrected code. Seems I made several mistakes’.
It was my intention to concade output from visa and split it into array and loop through elements until number is found.
I think it's working now.
Henrik I didn’t use your code, since I don’t send any data to balance. The balance is constantly sending by itself.
I tried to look at concated output, though.
Often "k" or "ks" is sent. But ONLY when I use labview. When I use my normal terminal test program, not written in LabVIEW I get only number. Weird.
Message 15 of 19
(1,858 Views)
Few things to note:
1. Since you are using the termination character to get a full message, there is no need for the whole concatination and search mess.  When you have a full message, you can extract the value and units out.
2. Are you sure the parity should be "None"?  Typically if you are only using 7 data bits, there is a parity used for that last bit in the byte.
3. I just realized that you are doing this whole thing just for 1 measurement.  Serial communications work best when you keep the port open and you constantly read the data coming from the instrument.  Just use a queue, notifier, user event, global variable, etc to send the read value to another loop who needs it.  By opening and closing the port constantly, you can get into some weird synchonization issues.
Which brings up another question: how quickly is the instrument sending the data?  Specifically, is there any gap between the messages?  I had to interface with an instrument that had no gap between messages and it was a PAIN to get the communications to work correctly, even with a terminal program like putty, due to some of the most bizare synchonization issues you could imagine.
Anyways, here is a greatly simplified snippet for what your code should really look like
Message 16 of 19
(1,854 Views)
I’m pretty sure that the parity is none. I tried all other parities.
I will try opening the port in start off the measurements end close it ending measurements.
I also considered that myself.
About the sending speed. Its fast ! Several pr. seconds. Tried to decrease that with no luck. Will try again.
As it is now I need “the whole concatination and search mess. “ as I see it. Often the output is like this “k LF 3102.2 LF”
(LF = linefeed)
Will return, hopefully tomorrow 🙂
Message 17 of 19
(1,842 Views)
Obs: forgot your code - Thanks !
Will try it when the instrument is free 🙂
Maybe I don't need the interation then
Message 18 of 19
(1,841 Views)
Seams that it’s when opening the port that things go wrong.
Approximately every fourth time.
I receive a “k” or nothing and it will stay like that until the port closes and opens again.
So if the error occurs I wont get any data before having opened and closed again.
Message 19 of 19
(1,802 Views)
