Triggering an acquisition using digital output
I want to use a DAQ Card 6062E to trigger an external device. This device has an input line
for triggering whose usual state is +5V, and the acquisition is triggered when the state is 0V.
I would like to know if I can use eg the DIG line VI to trigger the acquisition, or if the DAQ card can be damaged if a digital line is configured as an output while connected to a device whose input has a +5V potential. Thank you.
(3,182 Views)
It seems that the triggering line is for output or those voltages are required for trriggering (from an external source).
Try to measure the voltage on that line ...
(3,182 Views)
Sorry, I misunderstood the problem.
It depends on the trigger device configuration inside your instrument. Thus, the DAQ board can be damaged in some circumstances.
Usually, it is a TTL standard trigger and should be safe enough.
(3,182 Views)
Lab Viewer, Thanks for your help.
I will use a relay to trigger the acquisition.
Philippe
(3,182 Views)
Philippe;
The safer way to do that is by using the Dig Line of your DAQ board to switch a realy and that relay would switch the input trigger pin to ground.
Hope this helps.
Filipe
(3,182 Views)
Filipe, I think the use of a relay
is a simple and safe solution.
thank you for your excellent idea.
Philippe
(3,182 Views)
