07-30-2021 04:18 AM
I have NI USB 6363 DAQ Card.
In order to trigger it externally to start data acquisition, can anyone guide about width of that external TTL trigger pulse having amplitude of 5V?
There is no information about it in the data sheet.
08-05-2021 02:38 AM
Can you please share what have you tried so far?
08-05-2021 03:51 PM
It has a 100MHz base clock. If you are using that one, you should be fine with any trigger that is at least 10ns long.