08-18-2008 07:00 PM
08-18-2008 08:06 PM
08-19-2008 07:53 AM
08-20-2008
10:45 PM
- last edited on
07-23-2025
12:32 PM
by
Content Cleaner
Hi DSharp,
You are correct that the software will account for the CJC value which you set as constant when computing the returned temperature value from the voltage reading. One thing I would suggest trying would be to see if touching the thermocouple or otherwise providing a path to ground for the thermocouple makes the signal return back to a value within the valid range. When measuring a floating source such as a thermocouple in differential mode, it is necessary to provide a path from the instrumentation amplifier inputs to the instrumentation amplifier ground, which can be accomplished through the use of bias resistors being placed between the inputs and AIGND. For more information on bias resistors and properly connecting signals, please refer to the Field Wiring and Noise Considerations for Analog Signals tutorial. There is also more information about the problem you are seeing in "Appendix B: Troubleshooting" of the NI USB-621x User Manual (page B-1). On this page, you will see a topic titled "I am using my device in differential analog input ground-reference mode and I have connected a differential input signal, but my readings are random and drift rapidly. What is wrong?" which also discusses making proper connections. Hope this helps,
08-24-2008 06:36 PM