Synchronising the NI-9775

Solved!
Go to solution

Greetings all,

 

As the title suggests I'm looking to synchronise the NI-9774 module input with an AO card on a cRIO-9030.

I realise this is normally done by extending the FPGA I/O node, but doing this for the AI on the NI-9775 gives me zeroes from the data terminal. None of the module-specific examples seem to show using the I/O nodes for this so I'm not sure if this is even possible?

 

If not, could someone please suggest a method for synchronising the AI and AO?

 

Any help would be very appreciated.

0 Kudos
Message 1 of 4
(4,717 Views)

Hey Kieran,

 

I couldn't find a reference to the NI-9774 on the NI website, did you mean NI-9775?

 

As it says in this guide: Synchronizing Analog Input C Series Modules with NI-DAQmx in LabVIEW (Yes this is for DAQmx, but I'm going to refer to the module hardware):

"The Reference Clocked Module (NI 9775) cannot be in record mode when placed in the same task as a Delta Sigma Module" .... "The Reference Clocked Module (NI 9775) cannot be in record mode when placed in the same task as a SAR or Slow Sampled module."

 

I think this is what is causing you to get erroneous data.

 

  1. Share a master timebase source
  2. Start the synchronized acquisition
  3. Acquire:
    1. Acquire data from synchronised modules with the same data rate OR
    2. Acquire data from synchronized modules with different data rates

This is a good link to get started

 

0 Kudos
Message 2 of 4
(4,694 Views)

I did mean the NI- 9775 apologies.

 

Ah ok, does that mean the AI (9776) and my AO (9262) cannot share an FPGA I/O node?

 

I'm not sure the latter is SAR or Slow Sampled so I'm not sure why thats the case.

 

0 Kudos
Message 3 of 4
(4,690 Views)
Solution
Accepted by topic author KieranW

I actually found the solution to this;

 

In order to take data from the 9775 there needs to be a ModX/Start I/O node with a True value wired to it before the I/O data node. This way, it is possible to synchronise it with the AO terminal.

0 Kudos
Message 4 of 4
(4,666 Views)