 Konrado
			Konrado
			07-10-2012 05:17 AM
Hi there!
I'm using the ni driver package for ocean optics spectrometers with the USB2000+.
I want to take spectra as fast as every 20 ms, but it takes four times as long only to read out one single spectrum, as can be seen in the picture of the attached example vi.
Hopefully someone has any idea to acceleratate this task 🙂
PS: In a manual for triggering the spectrometer i've read you can trigger it as fast as every ~ 3ms - what's the use of that, if the data readout is that slow?
07-16-2012 10:13 AM
Hey, You are reading the "User Data" with the Property Node "USB Raw". There you should get two Numbers: hi(x): 0 or 128 and lo(x): 4118. lo(x) is the referencenumber of the usb2000+ and hi(x) is the speed I think 0 is the faster possibility. Is it possible to choose it on your device, a switch maybe? Otherwise you can try to change it inside the read-vi.
Did you try to ask Ocean-Optics, maybye they have an idea.
Gregor
 drjdpowell
			drjdpowell
			07-16-2012 10:31 AM - edited 07-16-2012 10:37 AM
I can't see in your code where you are setting your integration time. What is the default integration time? Is it 80ms?
07-16-2012 12:23 PM
If the integration time isn't set manually it's 6000 µs by default.
I've tested with shorter integration times too, but that has no influence.
We've also learned, that the execution-time of the vi varies from day to day. Execution times of 40 ms and 60 ms have been observed as well.
Can temperature differences of the spectrometer cause such variations?
 James_G1
			James_G1
			07-17-2012 11:44 AM
Can you tell if the device is being run as high speed of full speed?
07-19-2012 04:06 AM
I guess it's High Speed.
But anyways, the transfer of 2048 doubles (= Intensity values of the pixels) with full speed takes only
So that should be responsible for only 10 % of the time.
 James_G1
			James_G1
			07-19-2012 09:28 AM
You are quite right.
(2048 pixels per spectrum X 16 bits per pixel) / 12,000,000 bits per second = 2.73 milliseconds per spectrum.
Even if we throw LOTS of overhead in, this is not going to do it. I am thinking it must be something in your system. I gets speeds more like 4 ms per spectrum when I do a speed test with a USB2000+
James
07-19-2012 11:30 AM
Perhaps it's firmware related?
I have still the old(est?) firmware 1.00.
Which firmware do you use?
 James_G1
			James_G1
			07-19-2012 08:27 PM
Wow, this is some pretty old firmware. I did most of my testing with 3+ but I have been told that the older firmware is supposed to be faster than the newer firmware so I'm not really sure about that.