RS422 write buffer, read data format

Hi, I intend to read a distance value from a laser sensor using an RS-422 to USB converter. I have used the serial read LabVIEW example, but I am not getting the expected result. I am unsure about the write buffer string in VISA Write, and the data extracted from VISA Read is not in a readable format. Please suggest

 

 

0 Kudos
Message 1 of 4
(171 Views)

Hi Parthiban,

 


@Parthiban_SM wrote:

I am unsure about the write buffer string in VISA Write,

and the data extracted from VISA Read is not in a readable format.

Please suggest


  • For the VISAWrite there is a clear description in the manual about the possible commands.
  • There also is a description how multi-byte results (especially the 18bit distance value) is provided by the device: 3 bytes with their lower 6 bits holding the measurement data. I don't see any code where you attempt to handle those lower 6 bits of each of the 3 bytes…
  • Yes, the data is provided in "unreadable" format: switch the string indicator to hex display!!!
Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 2 of 4
(148 Views)

You have the right idea, trying to remove the first 2 bits from each byte, but it's not quite correct. Here is how to do it, note that you can apply boolean operators directly on integers:

Byte reconstruction.png

0 Kudos
Message 3 of 4
(129 Views)

Hi fuko,

 

is there a reason to answer with a crosspost from Reddit?

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 4 of 4