Morning!
It sounds to me like your hardware is getting confused, as it only falls over when two commands are sent back to back, and even that it works for one iteration but not two. That sounds to me like it takes the first MOV command, it's ok for the POS? command but that's followed instantly by another MOV then POS? which is when it's tripping up.
As a diagnostic, try putting a small bit of timing in the test program. Use a sequence structure with a 50ms delay and put it directly after the MOV command. This should give your instrument time to right itself before it receives the next command. You probably don't even need a full 50ms. Try playing around with different times and see if it makes a difference. Also try putting the timing in different parts of the program and see what happens.
If it's not that let me know and I'll start looking up data sheets!
Best of luck,
Sarah
AE | NI UK
02-18-2006 06:35 AM
I'll try your suggestion Sarah, but I thought that was the whole point of using RTS/CTS flow control: these lines are only asserted when the device is ready to receive data?
Anyway, I'll check it out when I'm back at work (I'm off next week).
Best wishes.
02-20-2006 08:34 AM
03-29-2006 08:07 AM
I'm not sure if anyone will spot this late addition to the thread, here goes anyway.
In essence the problem seems to be that the E-621 was receiving commands faster than it could handle them and falling over.
For example, with 0ms delay in the command sending loop the device typically falls down after a few loops; with a delay of 50ms in this loop it works fine.
Just to confuse matters more, i've installed a USB2/2. This connects the serial E-621 to my laptop's USB port (I've been doing all the development on the same machine).
Now with 0ms delay it works fine every time! The only explanation (apart from a bug) that I can think of is that there is "traslation time" within the USB2/2 that slows things down to a workable level.
Comments anyone?
03-30-2006 04:20 AM
Hi,
Like you say, I assume the new hardware and software layer have added an additional buffer to the data transfer which is allowing sufficient time for your E-621 to process each command before it receives the next one.
Originally the data was going straight to the OS then to the hardware.
Now it's being passed to the USB232 driver, which instructs the OS to write to the USB port which in turn writes to the E-621.
This additional time delay is what your instrument seems to require in order continually receive writes.