Pressure Transducer Scaling Certificate
Hello
My pressure transducer has the following calibration certificate:
pressure (PSIG)                 Unit Data (mVdc)
          0                             -0.275
     100                              49.824
      200                              99.774
      100                              49.844
         0                              -0.264
Balance     -0.275 mVdc
Sensitivity:  100.049 mVdc
Calibration Factors:
Sensitivity = 20.01 mV/V
the excitation voltage = 5 VDC  will give an output from 0-100 mV
I will use the DAQ assistance to make the pressure measurement.
the question is:  What should be the two electrical and two physical values (IN mV/V Units) that I need to enter in the DAQ assistance wizard when scaling the pressure transducer by the two points method. Again please specify the values in mV/V Units.
(2,614 Views)
Accepted by topic author wisjaf12
wisjaf12,
You can choose any of two points at your convenience.
For example: First Value: 49.824 Second Value: 99.774. (Electrical). First Value: 100 Second Value: 200. (Physical)
CarmenC.
(2,574 Views)
CarmenC
Thank you. and if I want the electrical values in mV/V I just need to devide them by 5V which is the excitation voltage. is that correct??
(2,563 Views)
