My pressure transducer has the following calibration certificate:
pressure (PSIG) Unit Data (mVdc)
--------------------- ---------------------------
0 -0.275
100 49.824
200 99.774
100 49.844
0 -0.264
Balance -0.275 mVdc
Sensitivity: 100.049 mVdc
Calibration Factors:
Sensitivity = 20.01 mV/V
the excitation voltage = 5 VDC will give an output from 0-100 mV
I will use the DAQ assistance to make the pressure measurement.
the question is: What should be the two electrical and two physical values (IN mV/V Units) that I need to enter in the DAQ assistance wizard when scaling the pressure transducer by the two points method. Again please specify the values in mV/V Units.
Thanks
Solved! Go to Solution.
03-25-2013 09:19 AM
wisjaf12,
You can choose any of two points at your convenience.
For example: First Value: 49.824 Second Value: 99.774. (Electrical). First Value: 100 Second Value: 200. (Physical)
CarmenC.
03-25-2013 01:39 PM
CarmenC
Thank you. and if I want the electrical values in mV/V I just need to devide them by 5V which is the excitation voltage. is that correct??