Good Morning,
I have been trying to calculate the accuracy of the DMM PXIe-4081 with the resistance 4-wire measurement , 7 1/2 digits resolution , 100 OHM range and the resistor the I am reading is 0.5 mOHM. The number that I am getting is blowing my mind and it does not make sense at all specially this should be the most accurate DMM in the industry. I am getting 1.2 mOHM. I am pretty sure that I am missing something with my calculations. Any help please
12-11-2019 11:10 AM
Did you look at this document? It should give you everything you need to know about this device:
https://www.ni.com/pdf/manuals/375254d.pdf
12-12-2019 07:48 AM
Yes sir,
That is why I feel that I am doing something wrong with my calculations somehow and I wanted a second look at the thing.
Please and thank you
12-13-2019 08:10 AM
Why are you not using the numbers provided in the specification from NI? Do you think they misjudged or miscalculated something?
12-13-2019 08:16 AM
Do you mean these numbers ?
These numbers are to calculate your accuracy right?
I am using them to calculate my accuracy while I am reading a 0.5 mohm with 100 ohm range
12-13-2019 08:42 AM - edited 12-13-2019 08:46 AM
Hi Ziad,
what do you expect when trying to measure a 0.5 mOhm (or 0.0005 Ohm) value with a 100 Ohm range? That's just 0.0005% (or 50ppm!) of the measurement range!
So what do you expect when the DUT signal is in the same range as the accuracy specs?
@ZiadAlkhoury wrote:
this should be the most accurate DMM in the industry.
Yes, that might be the case. But it's "just" a generic DMM - and apparently no device specifically made for measurement of very small resistance values.
There are other devices for this kind of measurement! Have you ever looked for "microohmmeters"?