NI-MAx same model interface
Hi all,
I have a question about NI MAX setting.
When communicating with GPIB, there is no problem connecting one signal generator but
when communicating with two signal generators of the same model, a strange string is queried in the attached picture as follows:
then, What is the problem with this setting?
Is it an error caused by the same instrument model?
I'll be waiting for the reply.
Thank you.
(1,665 Views)
Hello yhpark37.
Is the GPIB Address correct?
(1,630 Views)
Accepted by yhpark37
@hanamaru wrote:
Hello yhpark37.
I can answer that!  The two devices are configured with the same GPIB address.  Probably the factory default setting which is always the same for the same model for Agilant.
Each device on a GPIB bus must have a unique address.
You can change the GPIB address on the devices from the front panel of the instrument.  Just read the manual for instructions on navigating the configuration settings.
"Should be" isn't "Is" -Jay
(1,616 Views)
Thank you for your reply.
I changed the gpib address by referring to the instrument manual and detected two instruments.
(1,608 Views)
