â12-10-2014 02:34 PM - edited â12-10-2014 02:38 PM
I am trying to measure two analog inputs(NI-PXI-4461) using daq assist on Labiew 7.1. I want to run the same code on both inputs and comparing them together on a graph. I am measuring 2 accelerometers and the most important parameters are measuring the g rms and phase difference. I keep getting an error code -1907 "The sampling frequency of the input waveform should be greater than 0 (dt greater than 0)." I have set the sampling freq to 2k. I am not sure what I am doing wrong. Any help will greatly be appreciated.
SparkEE
Solved! Go to Solution.
â12-11-2014 09:33 AM
Hi SparkEE,
Playing around with your code with a simulated device, I noticed that the output from the DAQ Assistant gets lost when you use the 'Convert from dynamic data' block. I checked this by placing waveform graphs before and after the block. If you look at this block, there is a 'channel' input. I wired this to the iteration terminal of the for loop (so now it specifies channel 0 then 1) and this seems to have resolved the error. Can you give this a shot?
â12-11-2014 10:19 AM
AirFrantz, you are the bomb!!!