04-13-2010 03:26 PM
Solved! Go to Solution.
04-13-2010 03:48 PM
Yes, it is related to using a constant cold junction temperature value and the actual temperature of the junctions changing throughout the day.
Set your CJC to use the onboard thermistor rather than constant within your task.
04-14-2010 08:51 AM
I have tried changing the CJC source from "Constant" to "Built-In" and the following error message comes up:
"Measurements: CJC Source cannot be set to Built-In for the
specified thermocouple channel. The physical channel does
not support built-in CJC temperature sensor.
If your hardware contains a CJC temperature sensor on the
physical channel corresponding to the built-in CJC source,
make sure that the hardware configuration (including and accessories
and/or terminal blocks) is correct. Alternately, specify a different
CJC Source, or use hardware with a built in CJC temperature
sensor."
04-15-2010 04:56 PM
Hi Zawer,
Your SCXI-1303 has a high-accuracy thermistor cold-junction temperature sensor, so the "Built-In" setting should be supported. Try resetting the 1102 in Measurement and Automation Explorer and then restarting LabVIEW. Is this error still being thrown?
04-16-2010 01:53 PM
It works now. The problem was that the 1102 was not properly configured in Measurement and Automation Explorer. Thanks for the help.
04-16-2010 03:27 PM - edited 04-16-2010 03:28 PM
EEsRULEtheWORLD wrote:[...]
Try resetting the 1102 in Measurement and Automation Explorer and then restarting LabVIEW. Is this error still being thrown?
[...]
I reset my 1102 pragmatically every time my applications start. I have had a few issues where the temperature will be "stuck" at a strange value a little above ambient upon startup, but since I started resetting all the time - no problems. I think it also helps with initialization, especially when people hit abort rather than doing a proper shutdown the last time the program was used - leaving the 1102 in an unknown state.
The reset certainly doesn't hurt anything, and is a piece of cake with the DAQmx Reset Device vi.
edit: Check out the funny spell check change.