Hello all,
I am trying to use the example code provided in the LV drivers for Keithley DMM. I connected to the DMM with USB and MAX identifies it. But it seems to have a timeout error when I try running the example code Keithley DMM7510 Acquire Signal.vi.
The error seems to come from the IDN query however the IDN response from the DMM matches the string expression + SN and version. VISA Read happens when Query is turned off, occurs when the computer tries sending a *RST command to the DMM.
Attached is the LV drivers for the Keithley DMM that I am using, below is a picture of the error out and MAX config
Solved! Go to Solution.
01-17-2023 07:54 AM
Have you tried testing the commands with the VISA Test Panel in NI MAX.
*IDN?\n and *RST\n commands.
Do you get the same outcome?
01-17-2023 08:21 AM
Yes, *IDN?\n query -
1: Write Operation (*IDN?\n)
Return Count: 6 bytes
2: Read Operation
Return Count: 52 bytes
KEITHLEY\sINSTRUMENTS,MODEL\sDMM7512,XXXXXXXX,1.7.12b\n
and *RST\n write comes back no errors-
1: Write Operation (*RST\n)
Return Count: 5 bytes
01-17-2023 09:57 AM
On my side runs ok, I'm using VISA 20.0
I was able to reproduce the error when selecting another port other than the one at the instrument
Ill take a look if I can find something else
01-17-2023 10:05 AM
That is odd. If it matters I am using a 7512, however that is just 2 separate 7510's. I tried using Ethernet connection instead of USB and produced the same error. When accessing through the Virtual Front panel via entering the ip in edge I can access the dmm just fine, however I do need to type in a login and password. Could this be the issue?
01-17-2023 02:22 PM
I continued digging and found the solution. I connected via ethernet and accessed the virtual manual panel for the 7512. I went into the configuration setting and changed the command style from TSP to SCPI and now it works. Looks like it saves that setting on power off as well.
01-17-2023 03:18 PM
Is good to hear that, glad you found it!