Is it possible to decode IRIG-B signal using 6601 card?

We are preparing for a test that requires both time-varying data and quasi-steady data to be recorded and coordinated with a sort of timestamp.
We can feed the IRIG signal into one of our NI-4472  channels in one DAQ system, but we need to coordinate with another DAQ system which is
sampling at much lower rates and time-averaging the data (using a 6224 card and external devices).  We have a PCI-6601 counter-timer card available
but I only see examples and discussion of reading the IRIG-B signal with counters on the 6608 card.
Can I do the same thing with the 6601 card?
 
Thanks
 
0 Kudos
Message 1 of 2
(2,471 Views)
Dave,


Unfortunately, the 6608 is the only device which performs IRIG-B decoding. Provided you perform some sort of start synchronization, though, you should be able to effectively measure the time between samples with a buffered period measurement task to within the accuracy of the oscillator on the 6602. This should work just as well. Please let me know if you need any clarification.

Regards,
Ryan Verret
Product Marketing Engineer
Signal Generators
National Instruments
0 Kudos
Message 2 of 2